Dr. Jan Martens: MP-E and Mitutoyo lighting, risers, R5 crop, and camera settings

  • This topic is empty.
Viewing 1 post (of 1 total)
  • Author
    Posts
  • #40560
    marksmith
    Keymaster

    Dr. Jan Martens’ test conditions and questions

    Jan initially tested a glued specimen on a card with the MP-E at 5×, a rotational stand, and the 3D-printed dome. Because the attached lens diffuser appeared not to clear the dome opening, it was removed and the dome was turned backward. The resulting image was strongly yellow. The test required flash power of 1/4 +0.3 at ISO 1250, f/8, and 1/100 s.

    Jan’s complete email: setup, results, and questions

    Dear Mark,

    In theme with how things have been going (i.e. slowly), I finally have an update on the setup and some of the questions (in purple) that popped up as I was testing things. I hope the in-line images get sent through. Otherwise, the original images are available here. Password is: Mitutoyo.

    First off, thank you for offering help with the installation and configuration. The videos and manuals gave me all the information I needed. Some of the details are a little bit scattered but diligently watching and reading most of the things through made it click together.

    I was trying different ways of lighting a specimen glued on a card, using the rotational stand and the 3D-printed dome to diffuse light. I ended up turning the dome around so it was more in front of the specimen. Had to remove the diffuser on the MP-E as it would not fit through the hole at 5× magnification. Not an ideal setup or result:

    Images supplied: image006.jpg and image007.jpg.

    I now would not use the dome or the rotation stand and instead use the XY stand with the box and aim the flashes at the diffuser on the lens, as I did afterwards with the 7.5× Mitutoyo, see below.

    Few questions:

    Would you use any other ways of achieving even lighting on a glued specimen like this?

    In the setup above, I had to set the flashes to ¼ +0.3 with ISO 1250, f/8, 1/100 sec. The dome takes a lot of light, and so does the MP-E at 5×. Are these settings to be expected?

    When shooting with the 3D-printed diffusers/reflectors, which white-balance setting would you use? The only in-camera solution I can think of is to set a custom white balance with a grey card and refer to that when using the 3D-printed parts.

    Switching over to the microscope objective, I followed the normal setup as in your tutorial video, adding a layer of foam in the box to mount the pin horizontally. This works beautifully and at 1/16 the Godox flashes pulled through a 300-layer stress test with a short 200 ms hold time on the rail. They showed a temperature warning at the end but they did still flash consistently. Probably best to give them more time or less power.

    Just to be sure I am not missing something: is vignetting expected on the microscope objectives? The 70–200 mm is set to f/2.8, focus to infinity. I did not notice this in your video tutorial, but this could be a mirrorless vs. DSLR thing. It is not major, but then I know this is to be expected. I noticed it on all microscope objectives. Lastly, any improvement suggestions on the setup below?

    Kind regards,
    Jan

    Concise answers

    1. Other ways to achieve even lighting: Either stage approach Jan used can be acceptable; the essential point is getting the light path right. The gimbal is preferred whenever possible for image quality and lighting. Use the risers to preserve that position. Use the rack-and-pinion stage when the specimen is under 4 cm or when precision translation is required.
    2. High flash power and ISO: ISO should not exceed approximately 2000. The dark, underlit result is not because the system inherently needs 1/4 flash power; it is because the lens-mounted diffuser was missing. The flash heads must be aimed at that diffuser. The inside of the dome must be intensely illuminated, while the plastic is designed to block line-of-sight/direct light from reaching the specimen.
    3. White balance: Set a fixed 5200 K. The image is excessively warm because it is being photographed through the plastic rather than with the intended reflected light path.

    Jan’s second question set: Mitutoyo workflow

    With the Mitutoyo objective and foam-supported horizontal pin, Jan achieved a successful 300-layer test at 1/16 power and a 200 ms rail hold, with a temperature warning at the end. Her answers are summarized below.

    1. Vignetting: Yes. Keep the 70–200 mm at f/2.8 and focus set to infinity. When framing permits, step down one objective magnification and set the full-frame Canon R5 to 1.6× crop to remove the dark outer field.
    2. Objective diffuser and specimen placement: The diffuser currently shown is the wrong one for this objective configuration, which is why the image is washed out. Use the large-aperture diffuser with the large bellows or Lieberkühn reflector; it prevents stray light from entering the objective and washing out the image. The current diffuser is for working-distance flexibility when that flexibility is required. Position the specimen near the front edge of the box, but not so far forward that it enters the flash heads’ direct line of sight; leave space between the foam and the specimen.
    3. 300-layer flash temperature warning: A 300-layer sequence at 1/16 power and a 200 ms rail hold can reach the Godox thermal limit even when flashes remain consistent. Allow more time between frames or reduce flash power before the next long stack.

    Preferred stage and lighting hierarchy

    The gimbal configuration is always preferred when possible. It provides the best overall image quality and lighting control. Add one or two 1 cm adapter-clamp risers to raise the camera enough for the attached lens diffuser to pass through the dome opening, while keeping the gimbal in its normal orientation.

    The rack-and-pinion stage is the second-best option. It performs very well with samples under 4 cm and whenever precise translation is required. Use it when its positioning advantages outweigh the gimbal’s preferred lighting geometry.

    Conversation with Dr. Jan Martens: lighting, diffuser setup, crop mode, and camera settings

    This record captures the configuration guidance provided for an MP-E 5× and Mitutoyo Micro Kit workflow. It is intended to be searchable for other customers working through similar lighting, thermal, and field-of-view issues.

    MP-E / diffuser configuration

    Keep the lens-mounted diffuser attached. It can pass through the dome opening when the camera is raised using one or two 1 cm adapter-clamp risers. The small black square riser clamps with red handles provide the required height adjustment while retaining the correct gimbal configuration.

    Adapter clamp risers and lens footplates

    Do not turn the dome or gimbal backward. Instead, move the flash heads farther back, direct their light through the box/dome side windows onto the lens-mounted diffuser, and let the diffuser reflect the light back toward the specimen. Do not direct flash light onto the specimen.

    The intended lens-mounted diffuser workflow is shown in this setup video.

    Exposure, white balance, and heat

    • Start around 1/16 flash power; keep the working range roughly between 1/8 and 1/64, rather than 1/1 to 1/4.
    • Use a fixed 5200 K white balance, not Auto. A grey-card custom balance can be used to validate the setup, then keep the result fixed for the stack.
    • A 300-layer stack at 1/16 and a 200 ms hold can trigger flash temperature protection. Lower the power or increase the interval before the next frame.

    Vignetting and Canon R5 1.6× crop

    At f/2.8 and infinity, the 70–200 mm remains the recommended lens configuration with these objectives. The dark corners seen in the full-frame R5 view are expected in this optical combination. Where framing permits, use one lower objective magnification and enable the R5 1.6× crop:

    1. Press MENU.
    2. Open the red Shooting tab.
    3. Select Cropping/aspect ratio.
    4. Select 1.6× (crop), then press SET.

    The crop records the central APS-C-sized area, removes the dark outer field, and reduces the field of view. The camera-settings workbook identifies the R5 cropped output as 5088 × 3392.

    Camera settings workbook

    Download: Scalebar Files and Camera Settings.xlsx

    The workbook gives specimen size/field of view, focus step size, shutter speed, aperture, ISO, flash power, white balance, camera resolution, Zerene stack direction, and slider settings. For Mitutoyo work, use the Micro Kit rows for full-frame capture and the Micro Kit 1.6× Crop rows when the R5 crop is enabled. In the crop section the operating values remain f/2.8, 1/100 s, and 5200 K, while the R5 recording area changes to 5088 × 3392 and the field of view is reduced. For example, the 7.5× objective changes from a 4.8 mm field of view full-frame to approximately 3 mm at 1.6× crop, with a 20 µm step size and 1/16 flash power.

    Revised image captions

    5× MP-E beetle capture with a dark, yellow cast
    Image 1 — 5× MP-E dome test: This beetle result is too dark and strongly yellow because the dome is positioned too far backward/turned around and the correct lens-mounted diffuser has been removed. Much of the flash output is being blocked by, or transmitted through, the dome plastic instead of reaching the lens diffuser; the colour is not a normal white-balance variation. Raise the camera with one or two adapter-clamp risers so the diffuser fits through the opening, keep the gimbal/dome forward-facing, and aim the flashes through the side windows at the lens diffuser—never directly at the specimen. Use fixed 5200 K white balance and keep flash power in the 1/8–1/64 range.
    7.5× Mitutoyo beetle capture
    Image 2 — 7.5× Mitutoyo beetle: The diffuser shown here is the wrong type for this working distance. Use the large-aperture diffuser with the large bellows/Lieberkühn reflector so light is excluded from entering the objective and washing out the image. Reserve the small-bellows diffuser for 50×–100× objectives. The dark outer field is the expected vignetting of the f/2.8 configuration; use the Canon R5’s 1.6× crop and, if necessary, step down one objective magnification.
    Box and XY support configuration
    Image 3 — specimen position in the box: The specimen is flush against the back of the box, which is not the intended position. Pull the pin forward so the specimen sits near the front edge of the box, with a clear gap between the foam and specimen. This creates more working distance between diffuser and specimen while keeping the specimen out of the flashes’ direct line of sight.
    Camera, aluminium mounting bar, and flash mount setup
    Image 4 — low flash-mounting option: The included aluminium bar and flash mounts allow the flash heads to be mounted down low. The red circle in the reference below identifies that mounting area. See the low flash-mounting video for the setup.

    Aluminium bar and low flash mounting point circled in red

    Dome and flash head arrangement
    Image 5 — dome and gimbal configuration: Raise the camera by placing an adapter-clamp riser beneath the lens footplate, then fit the lens-mounted diffuser. The dome/gimbal is inverted here and will not work correctly in this orientation; it must face forward. Flash heads may be positioned wherever they can illuminate the inside of the dome, provided no direct line of sight from a flash reaches the specimen. The light should strike the lens-mounted diffuser and reflect back to the specimen; use 1/8–1/64 flash power.